RE: Calibration for TCP312A Probe
We recommend a 12 month calibration interval on most all of our instruments. This includes the TCP312A current probe. We do offer a service manual which covers performance verification and functional check that you can choose to utilize.
If you like you can contact service for a calibration quote.Below you can find the contact information for our Service Department and someone will be happy to help:
Tektronix/Keithley Repair/Calibration:
Web: https://www.tek.com/en/service-quote
Email: rmarequest@tektronix.com -
RE: ECCN for DMM6500
This information can be found on tek.com. Support - Export Codes
https://www.tek.com/en/support/export-codesDMM6500
ECCN = 3A992HTS = 9030320000
-
RE: What probes come with the TDS2104B?
You will receive one TPP0100 per channel.
-
RE: Accuracy of TDS3054B
We specify DC Gain Accuracy as of the reading. For example if you are looking at a 50mV reading you can expect it to be within ±1mV.
-
RE: Statement of volatility for DPO4014B
This information will be in our Declassification manual on tek.com.
link below.
https://www.tek.com/en/oscilloscope/mso4000-dpo4000-manual/mso4000b-and-dpo4000b-series-and-application-modules
Attached also. -
RE: Option 4-AFG on MSO54
The 4-AFG option enables the on-scope Arbitrary Function Generator. There is more information about the AFG in the primary user manual, starting on page 414: https://www.tek.com/en/manual/oscilloscope/456-series-mso-5-series-mso.
-
RE: Probes for MDO3014
I'm happy to help you find probes for your MDO3014.You need probes for a spectrum analyzer, and I'm assuming that you are referring to the built-in Spectrum analyzer on the MDO3014 unit. If so, Tektronix actually has an FAQ that covers this question, linked below.
You also may be interested in the TPA-N-VPI Adapter. This adapter accepts probes with BNC, TekProbe™, and TekVPI™ connection interface. If you are currently using a probe with one of these interfaces, this adapter would allow the probe to be used with N connector of the MDO3000.Links:
What probes are supported on the RF input of the MDO3000 and MDO4000 series?: https://www.tek.com/en/support/faqs/what-probes-are-supported-rf-input-mdo3000-and-mdo4000-series
TPA-N-VPI Probe Adapter: https://www.tek.com/en/manual/tpa-n-vpi -
RE: Looking for DAQ6510
We suggest the DAQ6510 with 7706 module. You'll need the KTTI-TSP option and custom script.
-
RE: Upgrading AWG5208 sample rate
Option AWG5200-8-2550 is the upgrade from 2.5GS/s to 5GS/s.
-
RE: P2220 Probe and MDO34
It will work. Just a manual set up to tell the scope that a 1x probe is connected and they will be good to go.