RE: Accuracy For AC Voltage of TBS2202B
Accuracy specs for the TBS2202B can be found in this document: https://www.tek.com/en/manual/oscilloscope/tbs2000b-series-oscilloscopes-tbs2000-basic-oscilloscope
-
RE: TCP0030A Accuracy Question
Hi,
The TCP0030A accuracy specs are as follows:
Accuracy, typical
DC: ±1% of readingDC to 60 Hz, ≤5 A: ±1%
60 Hz – 5 kHz, ≤5 A: ±1.5%
DC – 5 kHz, >5 A: ±1.5%
The % of reading implies an accuracy within a percentage of the measured value. For example, a DC signal at 1A would have a +- 1% accuracy, meaning it would be accurate within 10mA. That implies a range of 990mA to 1010mA for the true signal value. A 1A signal at 1kHz would have 1.5% accuracy, implying a range from 985mA to 1015mA.
-
RE: Accessory probe kit for DSA71254C?
Hello,
The DSA71254C scope does not include probes. Probes are seperate and selected at or after time of purchase. Any replacement parts for the probes would be found in the user manual for the specific probe.
-
RE: 5 Series MSO, TPP1000 accessories for high-temperature chambers
Thanks for contacting us,
The TPP1000 supports operating at temperatures up to 149 °F.
-
RE: P5200A acceptance test procedure
Hello:
Here is the link to the Instruction manual on tek.com:
https://www.tek.com/en/probes-and-accessories/high-voltage-differential-probes-manual-1The manual has Specification, Performance Verification and adjustment procedures for the P5200A, plus a lot more.
-
RE: P6139B or TPP0200?
Hello,
There will likely not be a large change in observed rise time using a 100MHz scope, since the scope is the limiting factor. The scope rise time should be around 3.5ns, while the TPP0200 probe is specified at 2.3ns typical.
The TPP0200 has 12pF input C, while the P6139B has 8pF. Depending on your system that may have a slight impact as well.
I hope that helps. Thank you!
-
RE: TBS2000B/MSO2024 AC Signal Measurement
Hello,
AC coupling is not available for triggering on those two models. Low Frequency Reject would be the closest option with these two scopes. The cutoff frequency for LF Rejection is 65kHz.
-
RE: MDO4034B-3 oscilloscope AC gain accuracy
Hello,
Unfortunately, we do not specify the AC gain accuracy for this scope. Our full list of specifications is in the Performance Verification manual: https://www.tek.com/en/oscilloscope/mso4000-dpo4000-manual/mso4000b-and-dpo4000b-series-0.
-
RE: TBS1032B 9V Circuit - Stopped Working
Hi,
Please power off, and remove power then power back on and do a default setup to verify that the the unit passes the Self Calibration. From the Utility button "Do Self Cal".
-
RE: Need to generate 9 V@100A DC
Hi, I recommended the 2260B-30-108