RE: Setting Lower Bandwidths with Generic BNC Probe
There are two methods of reducing bandwidth of a channel.
1. Using a MATH filter to limit the bandwidth.
2. Combining HIGH RES Mode with a reduction of Sample Rate which will also limit the bandwidth of the channel. -
RE: Setting Attenuation
Can you elaborate on your P6015A setup, the scope manufacture and model number you are using it on, and any other details you may have?
It is very important with this probe to have an earth ground, both at the probe body and at the interface of the probe to the scope (i.e., an earth-referenced scope channel).
I also recommend reading the full user manual for this probe. Due to the typical use case of this probe (high voltage), it is important to know its limitations and requirements. The manual can be found here: https://www.tek.com/en/high-voltage-probe-manual/p6015a
-
RE: Set scope settings from PC
Hello,
Unfortunately, there is not a tool that would help view and edit the settings files. The THS3000 is also not supported by TekVisa so it will not be able to respond to commands over the Talker/Listener.
-
RE: Set Max Amplitude
The 500 mVp-p to 1 Vp-p spec is for differential usage. When operating as single-ended, the voltage output is half of that or 250-500 mVp-p.
-
RE: Passive Probe not reading voltage
There is not any documentation regarding functional tests for the TPP1000 or TPP500 passive probes. Could it be that the probe is connected to a earth-isolated voltage source? Without a ground reference to the source you won't get a signal because the scope itself is earth grounded.
Depending on the bandwidth of your MSO54 you should have received 4 of either TPP probes. Do you believe one of the TPP probes to be damaged or defective?
-
RE: Service Manual for 2440 Oscilloscope
Thank you for contacting Tektronix. The 2445 went out of long term product support in 1995. Most manuals and parts are no longer available from Tektronix. Some customers have had success using one or more of these contacts:
--TekWiki
http://w140.com/tekwiki/wiki/--The Boat Anchor Manual Archive
http://bama.sbc.edu/ or The Boat Anchor Manual Archive Mirror: http://bama.edebris.com/manuals/--Bitsavers.org
http://bitsavers.org/pdf/tektronix/--Manoman.sqhill.com
http://manoman.sqhill.com/tek.php--Classic Tek
http://classictek.org/--Instrumentation and Metrology Labs
Phone: 877-357-5900 or 603-929-9990, Fax: 603-929-1329
Email: sales@imltest.com Web: www.imltest.com-- Manual Merchant: Linda Kaplan, P.O. Box 927792, San Diego, CA 92192-7792
Phone: 858 642-0785 Fax: 858 642-0885 Fax
Web: www.manualmerchant.com Email: linda_p@ix.netcom.com-- W7FG Vintage Manuals, 3300 Wayside Dr., Bartlesville, OK 74006
Phone: 800-807-6146,
Web: www.w7fg.com-- Manuals Plus, P.O. Box 549, TAD #601, Tooele, Utah 84074, Fax,
Phone: 1-800-345-4019 or 801-936-7000 Fax: 801-936-7004
Web www.manualsplus.com-- A G Tannenbaum, A. G. Tannenbaum P.O. Box 386 Ambler, PA 19002
Phone: (215) 657-0106 Fax: (215) 657-0520
Web www.agtannenbaum.com-- J Dosher Manuals (in UK):
Web: http://www.jdosher.pwp.blueyonder.co.uk/ email: jdosher@blueyonder.co.uk;-- Tom Gootee: email tomg@fullnet.com
Web: http://www.fullnet.com/u/tomg/manuals.htm--Bill & Stan's Tektronix Resource Site
Email: w7ni@easystreet.com
Web www.reprise.com/host/tektronix/home--Analog Services Inc
Phone: 270-276-5671 Fax: 270-276-5588
Web http://www.logwell.com/index.html--Big-List.com http://www.big-list.com/usedmanu.html
--Techrecovery used instruments
Phone: 877-837-8872
Web: http://www.techrecovery.com/-- Sphere Research Canadian Test Equipment
Web: http://www.sphere.bc.ca/test/ecg.html-- Consolidated Surplus
Web: www.consolidatedsurplus.comTektronix is not in a position to represent that parts or manuals provided by these companies will be complete and contain all updates. Nor can we represent that a particular manual will necessarily correspond to the customer's specific product because of different versions of a product. We are supplying these names merely to provide customers a possible source for manuals for obsolete products without any endorsement or recommendation.
-
RE: Accessories needed for Calibration?
There is no need to send the accessories. Please do not include them with your RMA.
-
RE: Semiconductor Sensor 4-Wire Measurement for DMM6500
Application requires Dry Circuit Resistance capability. DMM6500 does not have this feature. If you are looking for this capability, the DMM75100 is capable.
-
RE: Self-Calibration/ Updating Internal Table
Thanks for contacting us,
There is not a mechanism for self-calibration of the RSA306B. While you can follow the procedures in the Performance Verification manual (https://download.tek.com/manual/RSA306B-Real-Time-Spectrum-Analyzer-Specifications-Performance-Verification-077124301.pdf) to characterize the instrument and verify it is within spec, if the results indicate it needs to be recalibrated it will need to be sent to a calibration lab that supports it. You can find more detail on our calibration services webpage here: https://www.tek.com/en/services/calibration-services
-
RE: Seeing a "step" in the spectrum for the RSA306B
What you are seeing is the effect of a limitation of the RSA306B. The capture bandwidth is limited to 40MHz. This causes the RSA to capture several acquisitions and stitch them together, in your case DC to 30MHz, 30MHz to 70MHz, and 70MHz to 100MHz. This is normal behavior for the RSA, and unless you are taking measurements right at the noise floor it should not interfere with your measurements.