  • RE: Sleeve for P6139B Passive Probe

    Wed, 28 Sep 2022 22:33:08 GMT – Sleeve for P6139B Passive Probe

    I believe you are looking for the insulator sleeve, part number 204-1226-01. This can be seen in the user manual under Replacing the Probe Tip: https://www.tek.com/en/passive-probe-manual/p6139b-p5050b. Take a look at the image in the user manual and let me know if this is not the correct part.

  • RE: Best scope to perform bode plots

    Wed, 28 Sep 2022 22:33:06 GMT – Best scope to perform bode plots

    The 4, 5, and 6 Series scopes all feature built in measurements to perform bode plots. The measurement is labeled as frequency response analysis.

  • RE: Replacement for TDS3052?

    Wed, 28 Sep 2022 22:33:04 GMT – Replacement for TDS3052?

    We recommend the MDO3 Series as a replacement to the TDS3000 Series.
    If you have a list of specifications you need in for the new scope we would be happy to make a recommendation for the best scope for you.

  • RE: Using modes on the AFG31051

    Wed, 28 Sep 2022 22:33:03 GMT – Using modes on the AFG31051

    You will need to purchase the SEQ license, which enables the Sequence, Triggered, and Gated modes in Advanced Mode. The license name for the AFG31051 is AUP-AFG3SEQ-1.

  • RE: Electrometer Replacement

    Wed, 28 Sep 2022 22:33:01 GMT – Electrometer Replacement

    Recommending solution with 2290-xx power supply and 26xxB smu, PM's, adapters, cables. See notes.

  • RE: Measuring resistivity or conductivity for pellets

    Wed, 28 Sep 2022 22:32:59 GMT – Measuring resistivity or conductivity for pellets

    I recommended Delta Mode System 6220 + 2182A

  • RE: Error message on MSO64

    Wed, 28 Sep 2022 22:32:57 GMT – Error message on MSO64

    This message usually appears when the scope is turned off with probes still plugged in and then the probes are removed before being the scope is turned on again. The scope will always attempt to recall the last setup, including probe setup information.

    For the last part of your request, I am assuming that the message goes away after a few seconds and the scope stays on.

  • RE: Software for FCA3120

    Wed, 28 Sep 2022 22:32:55 GMT – Software for FCA3120

    TimeView is the recommended software for use with the FCA3120. Here is the datasheet for this software: https://www.tek.com/en/datasheet/modulation-domain-analysis-software. Here is the download for the software and you will get the software free for the first 30 days: https://www.tek.com/en/support/software/demo/timeview-modulation-domain-analysis-software-30-day.

    The soft case that is recommended for this product is no longer available. There is not a direct replacement for this product, but there is a similarly sized soft case that might work for you. This case is a little bigger than the other case and does not have the foam inserts that are in the other case.

  • RE: Need Hall Effect probes

    Wed, 28 Sep 2022 22:32:53 GMT – Need Hall Effect probes

    For Hall Effect probes, the highest we have is 500A RMS (750A Peak) with the TCP404XL: https://www.tek.com/en/datasheet/ac-dc-current-measurement-systems

    The only other option is the TRCP3000, which is AC only. It is a Rogowski architecture, not hall effect, so it cannot go down to DC. You can find the datasheet here: https://www.tek.com/en/datasheet/current-probes-0

    Thank you.

  • RE: Channel A, Current Offset on 2636B

    Wed, 28 Sep 2022 22:32:51 GMT – Channel A, Current Offset on 2636B

    Unfortunately it sounds like you'll need to send the instrument in for repair. Request an RMA here: https://www.tek.com/en/services/repair/rma-request

  • ‹